2.1Errorcode02SafetySoftware


The service codes 02.00.. to 02.22.. are recognised on the  CPU 68332 PCB; they
file:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB3F.tmp.pngcorrespond to  02.40.. to 02.62.. as recognised on the  Pneumatic Controller PCB.



Fault
Item
Cause/Remedy
02.00.001
si0_Err, New start, cause incorrect, reset
Monitoring of cold/warm start storage in RAM on  CPU 68332 PCB. Cause: Software error on  CPU 68332 PCB
02.00.002
si0_Err, Fast data chksum error, reset
Data packets between  CPU 68332
PCB and Pneumatic Controller PCB
cannot be transferred in full.  CPU
68332 PCB or Pneumatic Controller
PCB defective.
02.00.003
si0_Err, Safety software of partner CPU not started, reset
After starting, it was not possible within
10 seconds to establish synchronisation between  CPU 68332 PCB and Pneumatic Controller PCB. Cause:
CPU 68332 PCB or Pneumatic
Controller PCB defective.
02.00.004
si0_Err, Input data, chksum error, reset
Data packets between  CPU 68332
PCB and Pneumatic Controller PCB
cannot be transferred in full.  CPU
68332 PCB or Pneumatic Controller
PCB defective.
02.00.005
si0_Err, Vector for NMI
incorrect, reset
RAM or SW error on  CPU 68332 PCB
02.00.020
si0_Err, Si boot, case incorrect, reset
RAM or SW error on  CPU 68332 PCB
or Pneumatic Controller PCB.





























5664.500/800 Evita 4/Evita 2 dura        01.2002
Repair Instructions        – 4


file:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB4F.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB50.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB51.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB52.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB53.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB54.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB55.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB56.tmp.jpg



Fault
Item
Cause/Remedy
02.01.001
si1_ Err, Adjuster has incorrect pixel sum, X alarm, reset:
Screen content (with Evita 2 dura also 7- segment displays) is read out and compared to request. Cause: Hardware or SW error on Graphics Controller
PCB (Evita 4)/Front Panel PCB
(Evita 2 dura) or  CPU 68332 PCB or Pneumatic Controller PCB. If display plausible, then SW error; otherwise high probability of hardware error on Graphics Controller PCB (Evita 4) or Front Panel PCB (Evita 2 dura). Table for X:

1        Paw high adjuster selected
2        Paw high arrow selected
3        MV min adjuster selected
4        MV min arrow selected
5        O2 adjuster selected
6        Last adjuster plus one
7        Paw high adjuster not selected
8        Paw high arrow not selected
9        MV min adjuster not selected
10    MV min arrow not selected
11        O2 adjuster not selected
12        Paw high line 30 top
13        Pressure scale values top 30
14        Paw high line 60 top
15        Pressure scale values top 60
16        Paw high line 100 top
17        Pressure scale values top 100
18        Paw high line 30 bottom
19        Pressure scale values bottom 30
20   Paw high line 60 bottom
21        Pressure scale values bottom 60
22   Paw high line 100 bottom
23   Pressure scale values bottom 100
24        Paw high line 30 bargraph
25    Pressure scale values bargraph 30
26   Paw high line 60 bargraph
27    Pressure scale values bargraph 60
28   Paw high line 100 bargraph
29    Pressure scale values bargraph
100
30   Paw high display 1
31        Paw high display 2
32   MV min display 1
33   MV min display 2
34   Cold start screen







file:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB67.tmp.png5664.500/800 Evita 4/Evita 2 dura        01.2002
Repair Instructions        – 5


file:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB68.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB69.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB6A.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB6B.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB7B.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB7C.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB7D.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB7E.tmp.jpg



Fault
Item
Cause/Remedy
02.01.002
si1_Err, Paw high, Si copy, consistency X alarm, reset
Alarm values relevant to safety are checked for plausibility.
Cause: RAM or SW error on  CPU
68332 PCB or  Pneumatic Controller
PCB.
02.01.003
si1_Err, MV min, Si copy, consistency X alarm, reset
Alarm values relevant to safety are checked for plausibility.
Cause: RAM or SW error on  CPU
68332 PCB or Pneumatic Controller
PCB.
02.01.004
si1_Err, O2 adjust, Si copy, consistency X alarm, reset
Alarm values relevant to safety are checked for plausibility.
Cause: RAM or SW error on  CPU
68332 PCB or Pneumatic Controller
PCB.
02.01.005
si1_Err, Paw high, invalid value, reset
Alarm values relevant to safety are checked for plausibility.
Cause: RAM or SW error on  CPU
68332 PCB or Pneumatic Controller
PCB.
02.01.006
si1_Err, MV min, invalid value, reset
Alarm values relevant to safety are checked for plausibility.
Cause: RAM or SW error on  CPU
68332 PCB or Pneumatic Controller
PCB.
02.01.007
si1_Err, O2 adjustment, invalid value, reset
Alarm values relevant to safety are checked for plausibility.
Cause: RAM or SW error on  CPU
68332 PCB or Pneumatic Controller
PCB.
02.01.020
si1_Err, Si boot, case incorrect, reset
RAM or SW error on  CPU 68332 PCB
or  Pneumatic Controller PCB.
02.01.021
si1_Err, Xxx is set, case incorrect, reset
Only applies up to SW 1.04. RAM or SW error on  CPU 68332 PCB or Pneumatic Controller PCB
02.01.022
si1_Err, Control console, status, case incorrect, reset
RAM or SW error on  CPU 68332 PCB
or  Pneumatic Controller PCB.

















file:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB7F.tmp.png5664.500/800 Evita 4/Evita 2 dura        01.2002
Repair Instructions        – 6


file:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB80.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB91.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB92.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB93.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB94.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB95.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB96.tmp.pngfile:///C:\Users\mqs\AppData\Local\Temp\ksohtml\wpsEB97.tmp.jpg



Fault
Item
Cause/Remedy
02.02.001
si2_Err, Identifier has incorrect pixel sum, X alarm, reset
专业德尔格呼吸麻醉机维修
13414141187
Screen content (with Evita 2 dura also 7- segment displays) is read out and compared to request. Cause: Hardware or SW error on  Graphics Controller
PCB (Evita 4) /  Front Panel PCB (Evita
2 dura) or  CPU 68332 PCB or Pneumatic Controller PCB. If display plausible, then SW error; otherwise high probability of hardware error on Graphics Controller PCB (Evita 4) or
Front Panel PCB (Evita 2 dura). Refer to error 02.01.001 for X table.
02.02.002
si2_Err, Identifier not announced, X alarm, reset
Refer to error 02.02.001.
02.02.003
si2_Err, Value has incorrect pixel sum, X alarm, reset
Refer to error 02.02.001.
02.02.020
si2_Err, Si boot, case incorrect, reset
RAM or SW error on  CPU 68332 PCB
or  Pneumatic Controller PCB.
02.03.001
si3_Err, Keypad input sequence, counter incorrect, reset
SW or transmission error from
Graphics Controller PCB (Evita 4) or
Front Panel PCB (Evita 2 dura) via  CPU
68332 PCB to Pneumatic Controller PCB. Cause: Probably error on Graphics Controller PCB (Evita 4) or Front Panel PCB (Evita 2 dura).
02.03.002
si3_Err, Rotary knob button pressed, incorrectly displayed, reset
SW or hardware error when reading out keys on  Graphics Controller PCB
(Evita 4) or  Front Panel PCB
(Evita 2 dura).
02.03.003
si3_Err, Silence button pressed, incorrectly displayed, reset
Refer to error 02.03.002.
02.03.004
si3_Err, Reset button pressed, incorrectly displayed, reset
Refer to error 02.03.002.
02.03.005
si3_Err, Extraction button pressed, incorrectly displayed, reset
Refer to error 02.03.002.
02.03.006
si3_Err, Inspiration hold button pressed, incorrectly displayed, reset
Refer to error 02.03.002.
02.03.007
si3_Err, Expiration hold button pressed, incorrectly displayed, reset
Refer to error 02.03.002.

专业德尔格呼吸麻醉机维修、技术支持、配件供应!
13414141187





德尔格Evita4呼吸机故障代码.doc

德尔格Evita4呼吸机故障代码.doc

277 KB, 下载次数: 31